Normalisation of Data


dave9898

Recommended Posts

Hi all,

I am currently operating a piece of equipment that essentially collects particles and separates them based on their size. Essentially you have 8 stages, and each stage has a differing size of particles it collects. For example:

Stage----Size of Particles (D) (um)-----Mass Collected (M) (mg)
1-------------0.1 - 0.2------------------------------1
2-------------0.2-0.5------------------------------0.1
3-------------0.5 - 1--------------------------------1
4-------------1 - 2----------------------------------0.5
5-------------2 - 4-----------------------------------1
6-------------4 - 5----------------------------------0.5
7-------------5 - 9------------------------------------3
8-------------9 - 10----------------------------------2

Now, as you can see each stage has a different range of sizes collected. Stage 1 has a "width" of 0.1 um, while stage 7 has a width of 4 um. Because of these different stage widths it is common to normalise the plot to make the results indepedent of stage width, like so:

dM/dlogDp = dM/log(Du)-log(Dl)

where Du is the upper stage width and Dl is the lower stage width (ex, for stage one Du is 0.2 and Dl is 0.1). If you then plot this data:

Stage----------dM/dlogDp
1----------------3.321928
2----------------0.251294
3----------------3.321928
4----------------1.660964
5----------------3.321928
6----------------5.159426
7----------------11.75215
8----------------43.70869

Now, my question is, what exactly is the physical meaning of these new results. The initial results show me the mass of particles between 0.1 and 10 um collected after a certain time of experiment or some other factor. The total mass is 9.1 mg. But, these normalised results have much higher numbers. Do they have any inherent meaning other than being independent of stage width. Am I missing some fundamental calculus principle that imparts meaning to dM/dlogDp?

Thanks.

Link to comment
Share on other sites

I suspect that log is used in case it's not clear whether the important parameter is the linear dimension of the particle, the area, or the volume. But taking the log of the ratio of the sizes just means everything is off by a constant, so the relative comparisons still hold. (IOW, if the crucial parameter is the volume, everything is different by a factor of 3, but stage 1, 3 and 5 still give equal results.

 

But why you'd take that ratio is unclear to me. The particles in stage 1, 3 and 5 have unequal sizes, and the collected mass is equal. This analysis gives them equal weight, but you have to have collected a lot more particles in stage 1 to get that mass. (i.e. stage 5 could be a single particle with a mass of a mg, but stage 3 could be 10 particles each with a mass of 0.1 mg, and stage 1 could be 10,000 particles, each with a mass of 0.1 micrograms.)

 

So without more detail, it's difficult to say what the physical interpretation is.

Link to comment
Share on other sites

Just to give you a bit of background info, I essentially draw air through a device that separates out particulates based on their inertia and therefore size. There are 8 stages, each separates out the particles in the size range given in my initial post. Now, we know the mass of air that passes through the device (let's say 10 grams), and we know the mass of particles collected on each stage, so what I then do is calculated the amount of particles in a given air of mass (mgparticles/gair), for example:

Stage 1 mass collected = 1mg. 10 grams air drawn through. Therefore, 0.1mgparticles/1gramair. (Particles between 0.1-0.2 um)
Stage 2 mass collected = 0.1mg. 10 grams air drawn through. Therefore, 0.01mgparticles/1gramair. (Particles between 0.2-0.5 um)
etc etc.

Now, typically the 8 stages are summed and we get total particles (mg) per gram of air. This doesn't require normalisation. However, if we want to to see which size range of particles are predominant, then we need to take the size range of each stage into account. That's where we do the normalisation outlined previously.

I understand the principle of the normalisation. Indeed, how can you compare the mass collected in each stage when some stages collect a bigger range of particles than others?

What I can't get my head around is how the two are directly comparable. If you look at my original post the total mass collected is 9.1 mg, or 0.91mgparticles/gair if we assume 10g of air again. dM/dlogDp has the same units as mg or mgparticles/gair (you can do normalisation for either), but the numbers are far higher (72.5mg in fact).

So, ultimately, which one is correct? Are they directly comparable? Does the normalised have any physical meaning. Both each individual stage and the total mass of particles is much higher. If someone asked me "What mass of particles are in 10g of air?", would I tell them 9.1mg or 72.5mg?

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.